I have two INA219 Breakout Boards HW-831.
I connect Vcc to +5V and GND to GND.
SDA and SCL are connected.
V+ and V- are not connected.
I measured a voltage of 1.023V with a meter on V- to GND as well as on V+ to GND.
The readout via i2c Bus gives the same voltages.
I think the voltage should be read as 0V.
(The second device have the same result).
Can anyone help me.
INA219 1,023V on V-
Re: INA219 1,023V on V-
With V+ and V- unconnected, the results are undefined. What are the readings when you connect it to an actual circuit?
- ossi
When I connect a LiIon Cell with +Pin to V- and -Pin to GND.
The voltage on the cell is 4.04V (measured with a meter on the V- Pin to GND of the board).
I get a voltage of 4.012V reading the Register.
So I have to add 0x01c (d28) to the reading of the register to get the same voltage as on the V- Pin.
Next i connected my reference source of 1,0002V.
The readout of the register + the correction gives 1,000V.
I can measure a current of nearly 8uA when I connect the open V- to GND.
Is this in the tollerance of the INA219 ?
What is connected to V+? You can't measure current without a circuit and any current measurement made with one or more connections left floating will be undefined.I can measure a current of nearly 8uA when I connect the open V- to GND.
The resolution of the device depends on the internal gain setting as described in the documentation. With maximum gain, the resolution is 0.1mA.
https://learn.adafruit.com/adafruit-ina ... -2973528-5
A precision amplifier measures the voltage across the 0.1 ohm, 1% sense resistor. Since the amplifier maximum input difference is ±320mV this means it can measure up to ±3.2 Amps. With the internal 12 bit ADC, the resolution at ±3.2A range is 0.8mA. With the internal gain set at the minimum of div8, the max current is ±400mA and the resolution is 0.1mA. Advanced hackers can remove the 0.1 ohm current sense resistor and replace it with their own to change the range (say a 0.01 ohm to measure up 32 Amps with a resolution of 8mA)
- ossi
I have read the documentation.
But there is no hint that the V+ and V- can not be unconnected when I measure the bus-voltage.
There is an voltage outcomming of 1.023V at the V- Pin and this is the same as reading the voltage from the register.
Shorting this voltage to GND with a meter gives the current of ca. 8uA.
The final connection is:
V+ connected to GND with a 10 Ohm resistor (measuring the discharge of the cell).
V- connected to the + Pin of a LiIon cell
The measured current is OK
After considering the resistance of the lines, the measured value of the bus-voltage futhermore must be corrected by 28.
But thanks for your help that contributed to the understanding.
That is not what I said.But there is no hint that the V+ and V- can not be unconnected when I measure the bus-voltage.
* To measure voltage in any meaningful way, at least one of the inputs needs to be connected to something.
* To measure current in any meaningful way, both V+ and V- need to connected in series with your load. Measuring current without a circuit is meaningless.
- ossi
Everything you write is OK.
Thank you again for your information.
These have been helpful to me and excuse my bad english.