Adafruit Universal Thermocouple Amplifier MAX31856
by ericson95 on Wed Jul 21, 2021 7:32 pm
for my project I use a Adafruit Universal Thermocouple Amplifier MAX31856 (Breakout) with a Type T cable to measure temperatures in the area of liquid Nitrogen (-196 °C). The room temperature is shown correct (23 °C), but if I try to measure the temperature on a metal plate on the surface of the liquid Nitrogen I got a temperature round about -240 °C. This is not possible, because the normal temperature measure devise (which I used to compare the temperatures) shows -190 °C and the highest possible temperature could be not lower then -196°C (cause of the liquid Nitrogen). Also the description of this thermocouple says (on the Adafruit website) that this one has a measure area from "-210°C to +1800°C" with a precision of "output in 0.0078125° resolution". So what can I do?
Thank you in advance and best regards
Erik
ericson95
Re: Adafruit Universal Thermocouple Amplifier MAX31856
by mikeysklar on Thu Jul 22, 2021 6:31 pm
Hi Erik,
It sounds like you have things connected correctly and are getting a result outside our expected range for this amp. That is to be expected when you are so close the extreme end of the range. More here in the FAQ.
At extremely high or low temperatures, the measurements are not correct
The 31855 chip handles the linear range of the K-type thermocouples very well.  It does not provide correction for the non-linearities that occur at the extremes of the measurement range.  Thermocouple linearization for temperature extremes requires some curve fitting.  See this guide for more information and example code:
https://learn.adafruit.com/calibrating- ... earization
mikeysklar
by ericson95 on Fri Jul 23, 2021 1:16 pm
Hi Mikey,
thank you for your help.
Yesterday I tried the thermo sensor a second time and it showed me -174°C on the screen. This is also not the correct temperature. The thing is, I bought this one, because on the website they are saying it works good until -210 °C and my used temperature is -196°C... so normally it should work.
Is there a way to calibrate the thermo sensor?
Or is there a other good one from Adafruit, which I could use?
ericson95
by mikeysklar on Fri Jul 23, 2021 1:54 pm
We don't have a pre-calibrated ready to go unit that will work so close to the extremes. You will need to use the link in my previous reply to setup the linear handling and do your own confirmation calibrations.
mikeysklar
by ericson95 on Mon Jul 26, 2021 10:10 pm
Okay. Thank you very much...that looks really difficult for a beginner, is there a way to get help in coding? I use a Type T, is it possible to use the same script from the Type K?
ericson95
by mikeysklar on Tue Jul 27, 2021 12:48 pm
No the tables will be different for type-K vs type-T.
Since you know you are dealing with extremely low temperatures in a very specific range you can use these tables and write your own code to respond to voltage in the specific ranges you are work with.
https://www.thermocoupleinfo.com/type-t ... couple.htm
mikeysklar
by ericson95 on Mon Aug 02, 2021 12:05 am
Thank you Mike!
I will try out the two way calibration first and if it doesn't work I will try out this one.
Have a nice weekend!
ericson95