Hi I am using the SCD-30 - NDIR CO2 Temperature and Humidity Sensor to monitor a mushroom growing environment, the humidity is very high (90%) and I don't expect an un-coated component would last long.
I want to paint the sensor with a conformal coating but I don't know what areas to avoid so that the sensor continues to read CO2, humidity, and temperature.
Is there a drawing that shows areas I should avoid?
Thanks in advance
Conformal coating SCD-30
Moderators: adafruit_support_bill, adafruit
Please be positive and constructive with your questions and comments.
- adafruit_support_mike
- Posts: 67446
- Joined: Thu Feb 11, 2010 2:51 pm
Re: Conformal coating SCD-30
Take a look at the Design-in guidelines document:
https://sensirion.com/media/documents/8 ... nes_D1.pdf
It shows the temperature/humidity and CO2 openings on the first page.
https://sensirion.com/media/documents/8 ... nes_D1.pdf
It shows the temperature/humidity and CO2 openings on the first page.
- Mushrooom
- Posts: 8
- Joined: Thu Dec 29, 2022 3:30 pm
Re: Conformal coating SCD-30
Just wanted to update this post with what I learnt from applying Conformal coating to the SCD-30.
The sensor was in some way damaged or the calibration was thrown off by something about the process. Maybe the solvent fumes from the coating or maybe the black light I used to check coverage of the coating. I read the spec of the SCD-40, different sensor I know, but interestingly it points out how the solvent fumes can cause issues with that sensor. I am presuming the SCD30 is similar.
The conformal coating I used was - 422C Conformal Coating - Silicone with UV Indicator
The sensor was in some way damaged or the calibration was thrown off by something about the process. Maybe the solvent fumes from the coating or maybe the black light I used to check coverage of the coating. I read the spec of the SCD-40, different sensor I know, but interestingly it points out how the solvent fumes can cause issues with that sensor. I am presuming the SCD30 is similar.
The conformal coating I used was - 422C Conformal Coating - Silicone with UV Indicator
- adafruit_support_mike
- Posts: 67446
- Joined: Thu Feb 11, 2010 2:51 pm
Re: Conformal coating SCD-30
Try running the sensor in clean air for a few days.
You're right that the sensing path has probably picked up some VOCs. They should diffuse out over time.
You're right that the sensing path has probably picked up some VOCs. They should diffuse out over time.